Indeed. I have to say that I usually have a quick check of the IR and low-resistance ('continuity', if you must!) measurements of my ('uncalibrated') MFT whenever I use it, using my little 'resistor box' which always lives with it.realistically for an instrument like this "calibration" probablly consists of little more than sticking a few resistors on it and checking the results are in-tolerance. It's not exactly a precision instrument in the first place, nor does it need to be.(or, of course, get it calibrated!)
That's quite a high lower limit for many purposes. My Fluke MFT measures 'IR' right down to 10kΩ (which it displays as 0.01MΩ).The tester is only specced down to 3M (not 2M as I said in my original post).
I agree that what happened with 2.2MΩ could be a bit confusing but, other than that, what you report sounds consistent with a machine that has a 'lower limit' of 3MΩ.When I tried with a 1M resistor it gave a reading arround 1.8,beeped at me (it's supposed to beep below and showed a corresponding "buzzer" inducation on the display. A short circuit did much the same. Slightly more worrying was the 2M2 case, the beep and it's indicator on the display were still there and the decimal point were still there but a large number of digits on the display turned off leaving a display fairly similar to the over-range case (which shows a "1" on it's own with the decimal point and all the other digits turned off), which seems like a potentially confusing software bug, especially if you were using the meter in a noisy environment. ... A 10M resistor and a 3M3 resistor read fine.
Kind Regards, John