There seem to have been some new developments regarding my 'old' 1652, which I'll be writing about a bit later, once I have got my head around them!
This is far too confusing, inconsistent and variable for me to explain properly at present, but an initial update ...
Yesterday, after having observed that the presence of the "LEAd" message does not prevent a 1652 doing any tests for which the required leads are present, I felt that I had to check that I had not been totally silly all along, by not realising that my old machine would still work, despite the "LEAd" message appearing when I switched it on!
I therefore re-visited my old machine (still 'in pieces'), after what has probably been 3-4 weeks of 'neglect' (with batteries disconnected).
When I first switched it on, I thought that all was lost, because I repeatedly got the dreaded "Err1" message ('return to manufacturer). I then spent quite a long time playing with switching it on/off, with most/all possible combinations of leads and fuses present/absent. Somewhere along the line it suddenly started 'working' to at least some extent, and that has (so far) remained the case. However, all sorts of bizarre things are happening, and not all of them consistently. For example ...
When I switch it on with leads plugged in, it sometimes, but not always, momentarily shows the "LEAd" message, but that goes away. Just occasionally, it produces the "Err1" or "FUSE" message on power up, but that problem seems to go away if I just power down/up again.
The "LEAd" message has occasionally popped up briefly (but gone away spontaneously) whilst I've been using the machine.
Sometimes, when (after successfully switching the machine on) I connect the leads to 'the mains', it immediately shows the "FUSE" message, and refuses to do anything. It may be a 'coincidence' but, on several occasions I have got that problem to go away by bypassing the L fuse (the 'internal' one), despite the fact that the fuse is fine and the holder also seems fine. Wiggling the fuse makes no difference.
Prior to the recent illness, the machine was producing (per my 'testing box') pretty accurate results for all functions. However, now there seem to be all sorts of inaccuracies. It is under-reading voltage by about 5% (might not sound as much, but it means that, say, 240V is showing as about 228V). It is seemingly over-reading loop impedances by a bit over 5%, and low resistance ('continuity') measurements currently seem to be 'all over the place'.
Other than the measurement errors, little is particularly consistent, and most issues seem to be at least partially related to 'leads present' and 'fuse' tests. I've tried shaking/prodding etc. and can find no evidence of mechanical intermittency.
Given that the machine has partially recovered after a period of 'neglect', I wonder if some environmental factors may have been at play. It usually lives in my cellar, which has a very damp atmosphere, so maybe it would be worthwhile bringing it up into a more warm/dry environment for a bit.
If/when I have more to say, I'll report back!
Kind Regards, John